It looks like you are experiencing the exact same issue as has been discussed in this thread:
Briefly, probably some reagent have not reached the bottom of the well. We have seen these blank wells a couple of times, and we have yet to experience a blank well that cannot be "rescued" by repeating the detection steps again. Thus, the problem is not a lack of coated capture antibodies, but a one failed detection step. Please repeat the detection steps in the blank wells as follows:
1. Wash the blank wells 5x with 200ul PBS.
2. Add detection antibody at 1ug/ml in PBS containing 0.5% FCS and incubate for 2h.
3. Wash the blank wells 5x with 200ul PBS.
4. Add SA-ALP in PBS containing 0.5% FCS and incubate for 1h.
5. Wash the blank wells 5x with 200ul PBS.
6. Add substrate and incubate for 10min.
I would be surprised if you didn't get results looking like this:

So what to do to avoid blank wells in future experiments? Well, if you are too careful with emptying the pipette, there is a risk for a bubble formation in the bottom of the well, that blocks the following reagent to reach the bottom. Solution:
•Look carefully at the plate for odd looking wells after addition of liquid.
•Remove bubbles by tapping the plate against the bench.
•Do not touch inner wall of the well with the pipette tip.
Hope you manage to resolve the issue! Please come back to us if you don't (or, yeah, please do if you resolve it as well, would be nice for us to know :))