    Guest sarah2020-01-10T15:30:32Z

    Hi Mabtech,

    Have you tested how long the antibody-coated plates last for? Do you know if they would last up to 2-3 days/longer if kept in at 4oC, rather than just overnight? I need to do experiments that can be delayed (at short notice) up to a few days as it depends whether or not samples can arrive on specific days. Therefore, sometimes I will need to plate the capture antibodies 2-3 days in advance so I'm hoping that this won't affect the FluoroSpot assay?

    Thank you,

    Sarah

  • C
    Christian@mabtech.com2020-01-10T22:10:08Z
    Admin

    Hey Sarah,

    Yes, I have done this many times without any problems. One thing to consider is that the antibody coating solution inside of the wells will start to evaporate even when kept in a cold room with the lid on. So one can wrap the plate in Parafilm to minimize evaporation.

    Otherwise all liquid could evaporate and leave the membrane exposed to air. Not good.

    Also, any contaminant that entered into the plate during the plate coating process would get extra time to "grow". So be sure to etoh activate and add capture antibody in sterile conditions. But that is a given ofcourse.

  • GG
    Guest Guest2022-06-07T20:47:10Z

    We are using the FluoroSpot Path: Human IFN-g/IL-2 (SARS-CoV-2, S1scan+SNMO). The un-opened pre-coated plate was left in the box with the all other reagents, except the peptide pools, at 4-8C for 5 days by mistake. Is the plate still good to use?

  • C
    Christian@mabtech.com2022-06-07T20:50:55Z
    Admin

    1 minute ago, Guest Guest said:

    We are using the FluoroSpot Path: Human IFN-g/IL-2 (SARS-CoV-2, S1scan+SNMO). The un-opened pre-coated plate was left in the box with the all other reagents, except the peptide pools, at 4-8C for 5 days by mistake. Is the plate still good to use?

    Yes, should not be a problem. Take the plates out (keep them in aluminium coloured container) and let the unopened plates adjust to room temp before starting your next experiment.

  • GK
    Guest Kristina2022-06-07T20:52:11Z

    We are using the FluoroSpot Path: Human IFN-g/IL-2 (SARS-CoV-2, S1scan+SNMO). The un-opened pre-coated plate was left in the box with the all other reagents, except the peptide pools, at 4-8C for 5 days by mistake. Is the plate still good to use?

  • GK
    Guest Kristina2022-06-07T20:52:54Z

    Thank you!