Hi Sanda,
I think your standard curves look good, and the background levels are low. So the assays seem to have worked. The OD signal increases with development time, on both samples and standard, and I think 60 min is a good developing time for pNPP (30 min is a bit too short, why you shouldn't focus on those results). You have diluted the samples with a dilution factor of 2X and 3X, and 3X is often below the lowest standard point and you shouldn't use these measurements. Since you are analyzing plasma samples, you need to dilute the samples at least 2X in the ELISA diluent, so, therefore, you cannot dilute the samples less. The reason why you need to dilute the samples in ELISA diluent is that the diluent prevents false-positive read-outs which may be caused by interference of heterophilic antibodies found in serum and plasma.
My guess is that your samples have these very low levels of cytokines. I'm not an expert on PTSD but several factors might affect the levels, like timepoint for sampling, or disease progress etc? Also very important to use a diluent that prevents false-positive read-outs, I hope this has been done previously as well.
Best regards,
Lena